View on GitHub

IRAF Community Distribution

IRAF maintained by the community

Home | Installation | Packages | X11IRAF | PyRAF | Forum

SENSFUNC, CALIBRATE: Units and definition of the sensitivity function

Frank Valdes wrote on Aug 31, 1999

Q:  For the graph type `s' -- what exactly is the y-axis scale
(this is `sensitivity' vs wavelength. I would have guessed
that a total-instrument efficiency would have been a good
candidate, but the numbers don't look right.


A:  Below is the relevant section of the help page for SENSFUNC.  I admit
that the answer to your question is not completely obvious.  What you
want is from equations (1) and (2):

     S = 2.5 log (O / (T B F))

It is clear in the help for CALIBRATE, also given below, which is what
applies the sensitivity curve derived by SENSFUNC.


Cheers,
Frank Valdes


>From help for SENSFUNC:

    The calibration factor at each point is computed as
    
            (1) C = 2.5 log (O / (T B F)) + A E
    
    where O is the observed counts in a bandpass of  an  observation,  T
    is  the exposure time of the observation, B is the bandpass width, F
    is the flux per Angstrom at the bandpass for the  standard  star,  A
    is  the  airmass  of the observation, and E is the extinction at the
    bandpass.  Thus, C is the ratio  of  the  observed  count  rate  per
    Angstrom  corrected  to  some  extinction curve to the expected flux
    expressed in magnitudes.  The  goal  of  the  task  is  to  fit  the
    observations to the relation
    
            (2) C = S(W) + AE(W)
    
    where W is wavelength, S(W) is the sensitivity function, and E(W) is
    a residual extinction function relative to the  extinction  used  in
    (1).   In  later  discussion  we  will  also  refer  to the residual
    sensitivity which is defined by
    
            (3) R = C - S(W) - AE(W)
    
    The sensitivity function S(W) is output as an one dimensional  image
    much  like the spectra.  The sensitivities are in magnitude units to
    better  judge  the  variations  and  because  the  interpolation  is 
    smoother  in  the logarithmic space (mags = 2.5 log10[sensitivity]).


>From help for CALIBRATE:

    The sensitivity spectrum is in units of:
    
            2.5 * Log10 [counts/sec/Ang / ergs/cm2/sec/Ang].
    


----- End Included Message -----

Last post on Aug 31, 1999