Frank Valdes wrote on Oct 20, 1998

Q:  I am going to do some observing with NICMOS III spectrometer (at
Gurushikhar Telescope near Ahmadabad)in the J H K bands during 11-14
Oct and obtain spectra in these bands for some 60 F G and K type stars.
Do you have any reference or digital fluxes for standard stars in these
bands which we can install at the IRAF library at IUCAA so that after I
come back I can use the fluxes to do the flux calibration of the
program stars.


A:  You have reopened a discussion which involved a number of people, which
started with a question by Margaret Hanson about the existence of a
near-IR spectrophotometric library.  My answer, which was confirmed
by several others at CTIO and U Arizona, is that no one has done the
same sort of observations that were done in the optical by Phil
Massey and others to generate the standards one finds in the IRAF
ondestds directory.

There appear to be several reasons for this.

1.  It is a lot of hard work.  Phil elaborated for me the amount of
observing with the IRS and other instruments which went into the
list(s) he has published.  He cheerfully admitted that the issue of
telluric absorption lines, which is of great significance in the IR,
but not a particular problem in the optical, except at a few wavelengths,
was handled by simply choosing his bandpasses to miss any such
features.

2.  It can involve a lot of observing time.  Allocation committees are
generally not inclined, at least at KPNO, to give time to standards
or calibration programs on oversubscribed telescopes unless one can
make a compelling scientific case.  In the optical, one can often
provide strong arguments for the importance of combining spectra from
groundbased and spaceborne UV instruments. 


One approach to spectrophotometric calibration in the IR has been that
noted by Lazaro et al (1994, MNRAS, 269, 365).  They used the
absolute calibration data of Mountain et al (1985, A&A, 151, 399)
fitted to a Kurucz model atmosphere for Vega to obtain a hypothetical
Vega flux calibration spectrum.  They also note the absolute
calibration data of Cohen et al (1991, A. J., 104, 1650), but did not
employ their data.  Megessier (1995 , A&A, 296, 771) has raised
some questions about the accuracy of this technique (i.e., relying
on model atmospheres to establish flux calibration).  He refers to
a number of additional references and established an empirical
blackbody relationship, which he noted was not meant to imply anything
deep about the effective temperature of Vega.  You might want to look
through these references.

My one recent publication of IR spectra in a flux scale used the
calibration of Megessier, primarily because it was the most recent
and was at least a reference point for the fluxes.  I was forced to
use the published photometry for these objects, since my spectra
were not in any way calibrated due to slit losses.  My approach was
to convert the spectra, after dividing by a standard spectrum, to a
magnitude scale and scale the integrated magnitude over the wavelength
bandpass of the photometric filter to the published value.  This
is quickly described in A. J. 115, 2059, 1998.

It might be an interesting idea to put the Megessier calibration of
Vega into the IRAF library, but I am not sure how to do this.  My
perusal of the standards files show that the values must be in some
sort of strange units, since the "magnitude" of Vega varies from
1.191 at 3200 A, to slightly negative, and back to 0.507 at 8370 A.

Dick Joyce